lmh » Wed Jun 28, 2023 3:26 pm
An addendum: on a QExactive-type instrument, sensitivity is limited by how many ions you can collect in the C-trap without space-charging issues. That's the target. SIM and PRM only make sense if your analyte coelutes with something else.
If your analyte runs on its own, a large proportion of the ions in the C-trap will be your analyte, even in a normal full scan in which all ions are permitted into the instrument.
If your analyte coelutes with a huge contaminant, and you are using a mode that lets all ions into the instrument through the quadrupole, then most of the ions in the C-trap will be contaminant, and sensitivity will be reduced. In this case, if you use a SIM or PRM method where the quadrupole is selecting only the correct mass, you can filter out the contaminant and fill the C-trap efficiently with the correct mass, which will bring the sensitivity back up to the level you'd hope for from a clean peak.
But PRM will then suffer a bit of loss because only some of the precursor ions will fragment to the mass whose chromatogram you extract.
I hope I'm interpreting the inner workings of the QExactive correctly. Someone please tell me if I'm wrong! Incidentally, when I write SIM, I mean "proper SIM" using the quadrupole, not SIM as in SIM-Scan which they use for their normal scans. In the QExactive, that sort of SIM is just a scan across a very short range, and there's the option (advanced user settings) to set up quite a lot of them in a single event, but they are still short-range scans taken from one trap-ful of ions.