lmh » Wed Jun 28, 2023 3:05 pm
Assuming you're running a gradient method with reverse phase chromatography, you can test whether solvent contamination (aqueous, additives, and organic) is responsible for neatly-shaped proper peaks in a blank chromatogram. What can happen is that contaminants that aren't eluted by the starting conditions of the gradient accumulate on the column while you're equilibrating it (either before the first run, or between runs). When the gradient gets high enough for them to elute, they elute.
If you think this is going on, try a method where you equilibrate the column for three times longer than normal. In an ideal world, if you're accumulating contaminants, the contaminant peak will now be three times higher than normal!
You may have the data already. If the contaminant peaks are the same size in each run of a batch, except the first run, where they're much bigger, it probably means that you are in the habit of pre-equilibrating your column while you're setting up the instrument, and equilibrating it a bit longer than it's getting between normal runs!
But another common problem with a brand new system is that some item of the tubing is still leaching stuff/contaminated with stuff. I had a huge argument with a manufacturer once about the brand of LC-MS-grade isopropanol I'd supplied for the installation of an instrument. The engineer found contaminant peaks and complained mine was rubbish compared to theirs. He got some of theirs, and it still made peaks, and everything went rather quiet and sheepish. A year later, when the instrument came up for its first PM visit, a different engineer told me they'd worked out that the tubing they were using leaches contaminants into isopropanol for the first few weeks!
Of course if you're using an ancient column of uncertain history, it may be contributing all sorts of humps and bumps in the chromatogram itself.