Abysmal Precision
Posted: Tue Jun 27, 2023 9:39 am
Dear Reader,
I appear to be having some problems with the precision of a SIM method for the separation of 6 compounds + ISTD. The method I am using is as follows:
System – Agilent 7000d
Column – HP5 MS-UI, 30m, 0.25/0.25
Run Solvent – IPA
Injection Volume – 1ul – This equates to a vapour cloud of around 50% liner volume so backflash unlikely
Injection Parameters – 6 x prewashes (Solvent A and B), 3 sample washes, 4 sample pumps. No dwell time, draw at 300ul/min and dispense at 6000ul/min
Injection type – Splitless
Injector – Either 180C for 30s then ramp at 900C/min to 350C followed by purge at 1.33 mins or 180C isothermal – Former is slightly better overall
Flow (He) – 1.2ml/min
Oven – 50C hold for 3mins, ramp 15C/min to 150C hold for 2 mins, then ramp 40C/min to 265C hold for 8 mins.
Transfer Line – 250C (280 results in even poorer RSD!)
Source Temp – 230C
Quads – 150C
Overall, the chromatography is rock solid with great peak shapes, good separation and RT RSDs of 0.01% or less! However, peak area precision is poor and I have been trying to work through the problem with little success. For troubleshooting, I am injecting about 10PPM of each component and my within batch variation (5 sequential runs) varies from around 3% up to >8% depending upon the compound. Inter batch variation is pretty big with RSDs of up to 27% which obviously makes the method unworkable!
It seems to me that this is a problem that lies either with the ALS or with the injection, but am willing to be convinced otherwise. Any suggestions on dealing with this?
Kind Regards
TD2
I appear to be having some problems with the precision of a SIM method for the separation of 6 compounds + ISTD. The method I am using is as follows:
System – Agilent 7000d
Column – HP5 MS-UI, 30m, 0.25/0.25
Run Solvent – IPA
Injection Volume – 1ul – This equates to a vapour cloud of around 50% liner volume so backflash unlikely
Injection Parameters – 6 x prewashes (Solvent A and B), 3 sample washes, 4 sample pumps. No dwell time, draw at 300ul/min and dispense at 6000ul/min
Injection type – Splitless
Injector – Either 180C for 30s then ramp at 900C/min to 350C followed by purge at 1.33 mins or 180C isothermal – Former is slightly better overall
Flow (He) – 1.2ml/min
Oven – 50C hold for 3mins, ramp 15C/min to 150C hold for 2 mins, then ramp 40C/min to 265C hold for 8 mins.
Transfer Line – 250C (280 results in even poorer RSD!)
Source Temp – 230C
Quads – 150C
Overall, the chromatography is rock solid with great peak shapes, good separation and RT RSDs of 0.01% or less! However, peak area precision is poor and I have been trying to work through the problem with little success. For troubleshooting, I am injecting about 10PPM of each component and my within batch variation (5 sequential runs) varies from around 3% up to >8% depending upon the compound. Inter batch variation is pretty big with RSDs of up to 27% which obviously makes the method unworkable!
It seems to me that this is a problem that lies either with the ALS or with the injection, but am willing to be convinced otherwise. Any suggestions on dealing with this?
Kind Regards
TD2