Loss of MS signal

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hello! I have a problem with signal loss on MS Shimadzu. Lately I have had lower peaks until one day the signal was completely lost between two analyses. I cleaned the MS, did a tuning, and changed the capillaries leading from the LC and ESI capillary. After injecting the standard without the column, the signal is still lower than I would expect and after injecting with the column, all I see is noise.
Any idea what this could be, please?

The standards, MF are new, the column is also almost new and works well, cleaning and tuning was done as always.
Do you have a strong signal when tuning?

If so then it would point more towards the HPLC side of things, if tuning signal is low then I would focus on the MS side.

Possible bad electron multiplier? (or does Shimadzu use the PMT, I remember one manufacturer does but can remember which.)
The past is there to guide us into the future, not to dwell in.
2 posts Page 1 of 1

Who is online

In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry