I cannot make up my mind about validation, verification, revalidation, partial validation, full validation. I would be very grateful if anyone can explain the situation through a few cases:
1. What should I do if the HPLC method developed for the determination of paracetamol from a tablet containing only Paracetamol is applied for the determination of paracetamol from a tablet containing paracetamol + chlorpheniramine maleate (combined)?
2. The analysis of paracetamol in a tablet is performed by a validated HPLC method. What should I do if the HPLC is moved to the other laboratory ? Apart from the OQ/PQ of the instrument

3. Robustness studies have shown that a flow rate between 1 - 1.2 mL/min does not affect the quality of analysis. What should I do if the flow rate is changed to 0.5 mL/min to save solvent?
Thanks in advance.
I'm retired, did use cGMP for our pharmaceuticals. That said: I am NOT a cGMP or validation expert, so take that into consideration with my comments.
Case 1. I'd document that chlorpheniramine maleate did NOT interfere in the retention region of paracetamol using placebo tablets, and run a mini-study to show that placebo tablets spiked with paracetamol + chlorpheniramine maleate combined delivered correct results for paracetamol.
Case 2. A cGMP validation documents that the test method is valid, not that a certain HPLC unit it appropriate. I'd say do the OQ/PQ of the instrument and then do what your company does to document test method transfer, to demonstrate and document equivalent results.
Case 3. USP <621> and FDA ORA have guidelines about what parameters can be changed without requiring re-validation. I'm retired, and haven't looked up current regulations, but I'd guess that you'd need to go to a smaller-diameter column to use 0.5ml/min flow rate without re-validation.