We have 2 identical Agilent 1200 series HPLC's with a Quat Pump and VWD detector. Each is running the same method with the same column and mobile phase sources. Here is a brief description of the column, mobile phase, and method conditions:
Column: Agilent Eclipse XDB-C18
Mobile Phase A: 0.05% H3PO4 in Water
Mobile Phase B: Acetonitrile
Run time: 80min
Gradient Profile: 0-25min 66% A / 34% B
25-55min ramp to 10% A / 90% B
55-70min 10% A / 90% B
70-80min 66% A / 34% B
Problem: One instrument recently started giving ghost peaks and baseline weirdness on the ramp portion of the run and then lines out semi-clean during the 10/90 portion. The first analysis after the LC is started up for the day (after sitting overnight with 66/34 in the system) looks terrible (see image below). Subsequent runs are not near as bad. We found that if we flush the system with Acetonitrile for 10-15min during startup, the first blank analysis will be better but still not acceptable. Our 2nd LC with the same conditions continues to run fine, so I have been working on the assumption that something is contaminated in the problematic LC. During troubleshooting, I have tried many different things to clear this up, but it will not go away. Here is a list of what I have tried (not necessarily in order and somethings probably would not cause my problem, but I am grasping at straws):
- New mobile phases
- Mobile phase from a different water source
- Different brand of acetonitrile (all LC quality)
- Different H3PO4
- Flushed system with IPA
- Flushed degasser channels
- Swapped degasser
- Swapped MCGV (had a new spare)
- Replaced AIV cartridge
- Swapped AIV (had a new spare)
- Swapped pump module from another LC (not from the sister system to this one)
- Bypassed autosampler
- Bypassed column compartment
The first problem source that I went to was mobile phase, and I still think it has something to do with this. I do not have the knowledge to logically explain why this occurs. The longer the LC sits with the pump turned off, the worse the baseline will be during that first analysis. I think we could ease the problem by running a flush at startup, but we have ran this method on these 2 LC's for almost 15 years now (with regular preventative maintenance) without needing a startup flush.
I'm at the end of my limited expertise in solving this. If anyone has any thoughts or suggestions to share, I would greatly appreciate it.
Normal Blank Analysis

Problem Blank

Column: Agilent Eclipse XDB-C18
Mobile Phase A: 0.05% H3PO4 in Water
Mobile Phase B: Acetonitrile
Run time: 80min
Gradient Profile: 0-25min 66% A / 34% B
25-55min ramp to 10% A / 90% B
55-70min 10% A / 90% B
70-80min 66% A / 34% B
Problem: One instrument recently started giving ghost peaks and baseline weirdness on the ramp portion of the run and then lines out semi-clean during the 10/90 portion. The first analysis after the LC is started up for the day (after sitting overnight with 66/34 in the system) looks terrible (see image below). Subsequent runs are not near as bad. We found that if we flush the system with Acetonitrile for 10-15min during startup, the first blank analysis will be better but still not acceptable. Our 2nd LC with the same conditions continues to run fine, so I have been working on the assumption that something is contaminated in the problematic LC. During troubleshooting, I have tried many different things to clear this up, but it will not go away. Here is a list of what I have tried (not necessarily in order and somethings probably would not cause my problem, but I am grasping at straws):
- New mobile phases
- Mobile phase from a different water source
- Different brand of acetonitrile (all LC quality)
- Different H3PO4
- Flushed system with IPA
- Flushed degasser channels
- Swapped degasser
- Swapped MCGV (had a new spare)
- Replaced AIV cartridge
- Swapped AIV (had a new spare)
- Swapped pump module from another LC (not from the sister system to this one)
- Bypassed autosampler
- Bypassed column compartment
The first problem source that I went to was mobile phase, and I still think it has something to do with this. I do not have the knowledge to logically explain why this occurs. The longer the LC sits with the pump turned off, the worse the baseline will be during that first analysis. I think we could ease the problem by running a flush at startup, but we have ran this method on these 2 LC's for almost 15 years now (with regular preventative maintenance) without needing a startup flush.
I'm at the end of my limited expertise in solving this. If anyone has any thoughts or suggestions to share, I would greatly appreciate it.
Normal Blank Analysis

Problem Blank