I done validation tests (accuracy, precision....) and everything seems ok, except there is constant decreasing in retention time, especially for sulfates, in period of 3 weeks from 12.5 minutes to 11.8 minutes and still decreasing. Retention time for chloride was 3.6, now is 3.5, nitrate was 7 minutes now ist 6.8.
Pressure/flow rate is stable, what can be a problem? Is it that maybe concentration of KOH eluent (7.5 mM) is too low for this method (column)?