Can I still perform my measurements at 190 nm? I also use water as a mobile phase. Doesn't this take water off as a blank anyway?
Thank you!
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
A few more comments on this: I use water as a mobile phase. I use the same water to dissolve my sugar samples. Apart from the single sugar and the water, there are no other additives (no buffer or similar). Therefore, I thought that water is not visible in my chromatograms and therefore, the overlay does not cause a problem. Or do I have a mistake in thinking here?Getting an accurate signal of any compound at the wavelength 190 nm will be hard due to, as you noted, the absorption coming from your mobile phase (and anything else that may be in there, additives, buffers). ...
As for your comment about "taking the water off as a blank", I'm not sure exactly what this means? If you're assuming that when you inject a blank of water that the system will subtract that blank from your future analysis then that is not true.
You can. The sensitivity is poor in this case, but if the signal-to-noise ratio and the peak area precision are sufficiently high for your particular task (your particular concentration levels), you can perform the analysis. You can try to use slightly higher wavelength (e.g. 195 nm) and different bandwidth. However, usually sugars are detected with a refractive index detector.Can I still perform my measurements at 190 nm?
The method is quit simply. I only use water as mobile phase (no gradients, no more substances than one sugar and water). So I think it could work?If you could share more of your method parameters that could be useful,
^those detection modes are all more appropriate for sugars than UV.I rarely used low wavelengths for anything. But we had RID, ELSD, conductivity detectors too.

You meant amperometry, didn't you?^those detection modes are all more appropriate for sugars than UV.I rarely used low wavelengths for anything. But we had RID, ELSD, conductivity detectors too.