Having looked at your spectra, it seems to me that you have nothing to worry about!!
Thank you for your suppot !
However, I would like t osee the entire plots of m/z 58 vs m/z 59 across the acetone peak.
Here is entire peak plots for Acetone.
Also I have attached same for EthylAcetate
and ButylAcetate.
If we look at Acetone spectra, yes, 58 shifts to 43 after CH3 is eliminated and it leads to increase of 43 relative to 58. But I can't find explanation of absense of molecular peaks in case of EtAc and ButAc, may be as I am almost newbie in MG/MS
.
In your second chromatogram, the mass spectrum of acetone (from the library??) is not a good one because it does not show ANY m/z 59
This library spectra is taken from NIST library integrated into AMDIS, ver 2.73
2017. Time to time use Varian 2000 WS library - itseems good enough for my needs, but here also absence of molecular peaks creates problems with identification.
Have tried to decrease load by 10 times, but for mixtures of solvents I just loosing minor components, M+1 peaks decreases, but molecular peaks do not appear anyway....