Noisy baseline in DAD
Posted: Wed May 31, 2023 9:33 pm
by tidyup8460
I am fairly new to liquid chromatography and have been trying to set up the Agilent 1260 for a new method. Prior to setting up the Poroshell 120 EC-C18 column for sodium benzoate analysis, the column was flushed in its appropriate solvent for long term storage. The old mobile phase bottles were switched out for the new bottles of buffer (7.8 mM ammonium acetate buffer in acetonitrile pH 4.2 (72:28) and (15:85), degassed) for the new method. The system was flushed overnight to remove any salt buildup from the previous method with 10% isopropanol (column compartment heated to 60C), then rinsed with water. With the Poroshell column placed on the HPLC and the column temp off, the system was purged with the new mobile phase for at least an hour, then the purge valve was closed. The column was equilibrated with the new buffer, but the DAD baseline is still noisy (~5 mAU difference). Can anyone suggest what else I can do to reduce the baseline? I can provide more details if needed.
Re: Noisy baseline in DAD
Posted: Thu Jun 01, 2023 8:49 am
by H.Thomas
Is this random noise or do you see a regular pattern? Could be the pump then...
Posted: Thu Jun 01, 2023 2:52 pm
by Consumer Products Guy
My response is not directly related to the detector, but that seems like a complicated mobile phase for detection of sodium benzoate (as benzoic acid). I think if I was doing such assay, I'd simply use 0.5% phosphoric acid/water and ACN (or methanol).
What wavelength are you using?
Posted: Thu Jun 01, 2023 3:14 pm
by tidyup8460
It appears to be random noise. The scientist who worked on this had a clear baseline in their analysis. If it is a pump issue, how would I be able to tell?
Posted: Thu Jun 01, 2023 3:31 pm
by tidyup8460
This method was worked on by a different scientist before they left, and they had very clean baseline and clear separation in their analysis. The wavelength is set at 225 +/- 4nm, reference 360 +/- 20nm.
Posted: Thu Jun 01, 2023 5:22 pm
by vmu
ammonium acetate buffer... wavelength is set at 225 +/- 4nm, reference 360 +/- 20nm
Wrong combination. Acetate absorbs well at 225 nm; hence, the signal is noisy.
Doubtfully.