HILIC-ELSD organic compound separation

Posted: Mon May 08, 2023 11:01 am
by Kamyll
Hello everyone,

I would like to ask your advise in achieving peak separation between two organic compounds: ribulose 1,5-biphosphate (RUBP) and 3-phosphoglyceric acid (3-PGA) in HPLC using ELSD as detector. In the literature, they achieved peak separation between the two compounds by using the following mobile phases:

Mobile A : 20 mM ammonium acetate and 20 mM ammonium hydroxide in water (pH 9.45)
Mobile B : acetonitrile (ACN):water (85:15, v/v)

Unfortunately, we only have the Shimadzu GIST NH2 column whose working pH range is limited within 2 - 7.5, hence I am very limited in adjusting the pH of my mobile phases. I have tried using just 50 mM ammonium acetate as mobile phase A without adjusting the pH. Unfortunately, I could not manage to achieve peak separation despite using gradient elution program.

I would like to still explore different ways to make it work with our current column and only buy one if it is really necessary (last option). However, I have limited experience with chromatographic separation, let alone for similar organic compounds.

Would anyone have some helpful advice for this?

Re: HILIC-ELSD organic compound separation

Posted: Tue May 09, 2023 6:40 pm
by lylegordon
What is your gradient profile? What is the column specified for the method you are trying to use?

Unless your time/effort are abundant its probably cheaper to buy the column that you know will work vs designing a whole new method.

I'd expect a -3 vs -4 charge difference at pH 9.5 would be easy to separate here by your HILIC + anion exchange method, however you are around pH 7 without adjusting the pH of the mobile phase at all and the compounds could have a similar -3 charge state.

Have you tried running the separation at pH 3?