Thank you very much for putting me at ease on this! If I had the time I guess I could have run an area% and internal standard method then varied the column flow. I may have lost small peaks due to flattening putting the area/peak parameters below the integration parameters.None of the integration algorithms that I know of tied to any specific peak shape. The start of the peak is usually identified when the slope of the signal exceeds a certain trigger value, and the end of the peak is identified when returns near enough to zero the values in between are summed to give the area (yes, that's a gross oversimplification, but the point is that the details of peak shape don't matter).
What CAN matter with a distorted peak is a non-linear relationship between response and analyte concentration, but as long as the entire peak stays within the linear range of the detector, the shark-fin should not have a significant impact on integration.