Calibration Valve 5973 always open?

Posted: Wed Apr 12, 2023 8:26 am
by mvlo
Dear fellow chromatographers,

our MS System (Agilent 5973) shows steady peaks of the calibrant PFTBA during all acquisitions. If the valve is switched off under manual tuning the ion peaks do not disappear and change in abundance. Also if the valve is getting purged there is no click sound. Did anybody experienced something like this? Could it be that the valve is steadily open and can't manage to close?

I would be very grateful for any advice!

Re: Calibration Valve 5973 always open?

Posted: Wed Apr 12, 2023 12:44 pm
by mathieu_stability
hi mvlo

sounds like the valve / solenoid is permanently open.
either the valve is defective or the control voltage to the solenoid is continuously on.
you could try to unplug the solenoid connector which is behind the MS lower panel. this will tell you if the valve closes when no voltage is present.

Re: Calibration Valve 5973 always open?

Posted: Wed Apr 12, 2023 7:22 pm
by James_Ball
I agree. If it goes away when unplugged then there is a problem with the control board, if it stays then the valve is stuck open and probably needs to be replaced.

I have had them fail before, both open and closed.