Can't get Acetone to react with DNPH
Posted: Mon Mar 20, 2023 8:57 pm
by Pear
I have a method for aldehydes and ketones using DNPH with HPLC-DAD. Our formaldehyde method works great, but I can't seem to get Acetone to work.
I have tried 1-20ppm curves and also 20-50ppm curves. The peak area that shows up for Acetone increases slightly with increasing concentration, but it's nowhere near linear (1ppm gave 7 area and 20ppm gave 22).
My mobile phase is 60% Water : 40% Acetonitrile using a C18 column.
Reaction conditions: 1ml DNPH in 50ml acetontrile to give stock solution. 2.5ml of stock solution + 2.5ml acetone standard reacted for 30 minutes at room temperature.
Does anybody have any experience with Acetone detection with HPLC? All of the literature I can find says acetone reacts readily with DNPH.
Re: Can't get Acetone to react with DNPH
Posted: Tue Mar 21, 2023 2:23 am
by lylegordon
Can you just monitor acetone directly? It absorbs more than water/acn?
Posted: Tue Mar 21, 2023 2:14 pm
by Consumer Products Guy
Could the acetone be reacting and just not able to be separated from the excess DNPH ???
Posted: Fri Mar 24, 2023 4:20 am
by wss
The larger aldehydes and ketones often react much more slowly with DNPH than formaldehyde does. Try the same derivatisation but extend the reaction time out to a few hours, or try heating it up to around 40-50 degrees C.
Posted: Sat Mar 25, 2023 2:49 pm
by JI2002
Acidic conditions are required for derivatization of ketones and aldehydes if I remember correctly.
Posted: Wed Mar 29, 2023 9:53 am
by Gaetan Glauser
Indeed, IIRC we used trichloroacetic acid with DNPH.