Acetonitrile stability as a mobile phase
Posted: Sat Mar 18, 2023 6:08 pm
by quantonium
according to the "best practices" document by Agilent (https://www.agilent.com/cs/library/user ... ice_en.pdf), special care should be taken with acetonitrile as a mobile phase, but I can't seem to find any more information on this to confirm it. The document mentions how ACN should be kept in brown bottles to avoid photochemical reactions (I've also heard acids like formic acid degrades it?), as well as the need to monthly wash the LC system with 1 L of hot water to remove ACN reaction products. Is this common practice?
Also, as a bonus question: I'm very confused by the differing opinions on how long mobile phases last after mixing, with regards to microbial growth. It seems most agree that pure water with buffer at around pH 7 should ideally be changed daily. However, how long does water last with 0.1% formic acid? What about a 50H2O/50MeOH mix without acid?
In advance, thank you! (And thanks for all I've learned from browsing this forum for years!)
Re: Acetonitrile stability as a mobile phase
Posted: Sun Mar 19, 2023 2:45 pm
by Consumer Products Guy
We used 100% acetonitrile as one of our mobile phase components, and never used any special precautions; we would simply refill the reservoir when it got low.
As to aqueous mobile phases: sometimes we just 100% water as the aqueous fraction, often we used water acidified with acetic acid or phosphoric acid. With 100% water we switched out more often.
When we used THF or a mobile phase containing THF, we used amber reservoirs and only made up as much as was needed that day. We used THF without preservative.
Disclaimer: I only used HP/Agilent 1084B, 1050, 1100, and 1200 units; I felt there was no reason to look into other vendors as our reliability with these units was so great.
Disclaimer 2: I'm not necessarily recommending any of the above practices, just detailing what I did for 35 years....
Posted: Sun Mar 19, 2023 10:38 pm
by per_oxid
I have no clear answer but this are what I would guess, no actual data to prove it.
I think the question of 100% water as mobile phase and microorganisms was adressed well by the "trust you science" channel on youtube episode7. One day only. (0.1% FA i would assume not making much difference.)
I try to avoid 100% water when i can but then how long can i use a mobilephase with say 20% MeOH and 80% H2O? I dont know for sure but for mobilephase i made a guess of 2 weeks and for seal wash I accept 4 weeks, reasonable guess? If it is 50% MeOH i would guess it last a very long time probably 2 months. It is very practical to be able to use one batch of mobilephase for at least 1week of samples.
Volatile buffers can evaporate quite quickly so how good seal the bottle have makes big difference, investing in good better sealing lc caps is worth it. On my ms system the signal goes down over time as the buffer evaporates making ionisation less efficient. If bottle is in lc then i run no more than a week as caps are not sealed but if it is high in ACN % and sealed bottle cap and bottle is full then it can last many weeks, that is my guess and what I do.
ACN has been said to case sticky valves and a good practice is to switch a and b solvent lines to clean pumps from time to time as they are not cleaned by the gradient.
Posted: Mon Mar 20, 2023 5:09 am
by quantonium
Thanks Consumer Products Guy and per_oxid, good to hear what your practices are. I also watched the "trust your science" episode, and obviously got spooked :)
Posted: Mon Mar 20, 2023 12:36 pm
by DR
ACN, when left in a system long enough, can cause "sticking" check valves.
I've been told that it tends to react on the surface of the ruby balls commonly used along with sapphire seats in check valves to form polymers and that this is what results in sticking, floating etc.
Switching to ceramic check valves has helped a lot with this (typically in Waters Alliance systems).
Aqueous solutions allow growth of things you do not want in your HPLC systems, solutions containing at least 5-10% organic (ACN/MeOH etc.) tend not to allow such growth, but also tend not to remain at 5-10% organic if left long enough too.
The above guidelines have been pretty common to the several places I've worked over the years and try to strike a balance between absolute prevention of system issues and inconvenience to the people operating them.
Also remember that your best, most reproducible gradient methods usually go from 100%A to 100%B, so do the math and premix those mobile phases to allow for this once you're beyond the development stage for your new methods. Premixing also provides the advantages of having that little bit of organic in most A phases and/or diluting your THF so that you need not worry about things like bug growth and peroxide formation for at least a few days :).
Posted: Tue Mar 21, 2023 5:04 am
by quantonium
Sounds good. I'll likely add some methanol to the A solvent when I'm done with method development, then. I'm also considering doing the "hot water flush" as described by Agilent, as acetonitrile has been used a lot with this pump.
Posted: Fri Mar 31, 2023 1:44 pm
by DavidHPLC
For my experience:
Aqueous phases: Up to 3 days
Organic phase (and mixtures): Up to 1 month
We've had problems with system failure when gradient was run at 100% B (pure ACN) and those seemed to went away when reprogramming to 95%B.
"ACN makes sticky check valves" is what I was told once but to be fair I was told told that for MeOH when used with Alliance systems.
Posted: Fri Mar 31, 2023 5:45 pm
by James_Ball
One thing I have found is that if there is always flow then growth within the system is less of a problem. I like to use a very small continuous flow, 0.01 to 0.2 ml/min when in standby, unless it will not be used for a week or more, then it is better to flow 50%/50% aqueous/organic through the system before stopping the pump.
Of course this does not slow growth in the bottle for the aqueous, those need to be changed out often. Keep filters on the inlets to help remove any algae that could be pulled into the lines too.
Posted: Fri Mar 31, 2023 5:51 pm
by quantonium
Thanks for sharing, David and James. Some people at my group leave 20 µL/min running for some days, as you say. As of now, I do an end-of-day wash with 50% MeOH.
Posted: Sun Apr 02, 2023 8:48 am
by pksik
Some more information in patent document:
https://patents.google.com/patent/US4862907A/en
Not sure about polymerization inducing by ruby.
And articles:
https://www.chromatographyonline.com/vi ... onitrile-0
This is not in the HPLC pressure range but still interesting one :)
Polymerization of Acetonitrile via a Hydrogen Transfer Reaction
from CH3 to CN under Extreme Conditions
PS
Posted: Wed Apr 05, 2023 10:04 pm
by wss
Maybe we're just lucky (touch wood) but in my current lab we've never had any issues with algal, etc. growth in our aqueous phases, even pure deionised water sitting around for months. The best practice I tell my lab techs is to make up aqueous mobile phase fresh if it's cheap enough and keep it for a week if it's got a lot of the good (expensive) ammonium acetate in it. But there have been a few rush jobs where we've loaded up some old mobile phase that has been sitting around (sealed Schott bottle) for a month and thankfully had no issues.
The brown bottles for acetonitrile is also a good idea. Anecdotally however, we use 0.05% formic acid in acetonitrile as one of our mobile phases, which is stored in a colourless glass solvent bottle, and we only top it up once every couple of months and have had no issues with stability at least in the context of running successful analyses or causing instrument issues...
Posted: Thu Apr 06, 2023 2:23 pm
by DylanAnderson
I can attest to running pure acetonitrile through a system and letting it sit will indeed lead to sticking check valves. We use the sapphire seated valves and I've had to take them off and wash them a couple time last year. I also noted particulates accumulating in my bottle of 90% acetonitrile even when capped and stored in the dark over a duration of a couple months so that's still a mystery. I've started rinsing with more MeOH solutions and less ACN and my system has been running well.
Posted: Thu Apr 06, 2023 2:33 pm
Editing first sentence:
"We used 100% acetonitrile as one of our mobile phase components ON AT LEAST 5 HPLC SYSTEMS, and never used any special precautions; we would simply refill the reservoir when it got low.
Posted: Fri Jul 14, 2023 3:29 am
by mzchem
We use acetonitrile extensively in methods in our lab. We always purge the system with 1:1 ipa:h2o afterwards and keep it packed with it when not in use.
We have run into the sticky valves and other issues before, but the folks who came before me found that keeping some small percentage of water in the mobile phase containing high percentage of acetonitrile (when compatible with the method) combined with purging with ipa:h2o kept the issues to a minimum.
The professor who taught me the lc basics (who spent years working at Pfizer doing lc) always included some small percentage of the organic phase in the - normally - aqueous phase in order to keep the potential growth at bay.
We would keep an eye on our HILIC solutions with high acetonitrile content and watch for any particulate. You’d see it on occasion but not consistently or with any pattern.
Posted: Fri Jul 14, 2023 12:44 pm
by DR
Sticky valves and no suitable means to cut your ACN below ~90%? Switch to ceramic check valves. They're less prone to getting sticky.
