Page 1 of 1

ELSD signal wont show up in chemstation offline analysis

Posted: Fri Mar 10, 2023 12:49 pm
by Colloidchrom
Hi everyone

We recently coupled a Sedere LT 90 ELSD to our 1200 Agilent system. We also installed a driver that is compatible with chemstation and it seems to be working fine. The signal (voltage) is shown on the online spectrum in mV and we are able to run and observe the run online. However, when the report is generated and the file is transferred to chemstation offline, only the DAD1 signal is recorded and can be analysed.

Any ideas how to fix this? Greatly appreciated. In signal preferences, I have already added both the ELSD signal and DAD.

Re: ELSD signal wont show up in chemstation offline analysis

Posted: Fri Mar 10, 2023 4:59 pm
by Consumer Products Guy
I'm retired, but remember a similar situation from way back. We could see such external signal on the monitor but the signal was not being stored. The solution was to run a 9-pin remote cable from the 35900 to the computer.

Re: ELSD signal wont show up in chemstation offline analysis

Posted: Mon Mar 13, 2023 12:51 pm
by Colloidchrom
Thanks, CPG! That was it. Much appreciated.

Re: ELSD signal wont show up in chemstation offline analysis

Posted: Mon Mar 13, 2023 3:20 pm
by Consumer Products Guy
Thanks, CPG! That was it. Much appreciated.
Winner, winner, chicken dinner !!!!!

The blind squirrel finds a nut every once in a while......