Cleaning the UV/Vis detector measuring cell

Posted: Mon Feb 20, 2023 2:19 pm
by clinephd
Hello,
the measuring cell of my UV/Vis detector is very dirty, so hardly any energy gets through. I have already carried out a measuring procedure with nitric acid, water, acetone and acetonitrile. Minimal success could be observed. Do you know of any other cleaning procedures I could try?
(I changed the deuterium lamp on Friday).
Many thanks and best regards
:-)

Re: Cleaning the UV/Vis detector measuring cell

Posted: Mon Feb 20, 2023 4:13 pm
by tom jupille
After long exposure to intense UV, detector cells windows will "solarize" (often ending up looking like amethysts). Time to change the windows.

Re: Cleaning the UV/Vis detector measuring cell

Posted: Tue Feb 21, 2023 12:21 am
by pigsy1983
which brand is the UV detector, how many is energy of ref cell .