I am searching for an answer to the column sequence for a GPC instrument. I have a PL aquagel-OH 30 8mm column and a PL aquagel-OH 20 5mm column that will be placed in sequence. Which should go first and why?
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
The general rule is to put the packing with the largest pore size first. It may not be that terribly relevant for your analysis, but the idea is to dilute the larger MW as soon as possible, since the larger MW has the highest contribution to the sample viscosity.
2 posts
Page 1 of 1
Who is online
In total there are 29 users online :: 2 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], John Guajardo and 27 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], John Guajardo and 27 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science