I realize its not ideal for vitamins but its what I had installed. While out of the room the instrument went over pressure. When i tried turning pump back on it immediately went over pressure again. Using an Agilent 1100, binary pumpi G1312A
1st I removed column and left guard column on. pressure was ~50 at 1ml/min
2nd i tried column only and pressure rose quickly. I tried back flushing at starting at 0.1 ml/min and worked up to 0.5ml /min, pressure went to 400.
3rd removed column and opened purge valve, pressure did not go down. Pressure is now going down about 1 bar per minute. I removed purge valve to check gold seal and it looked new, was just recently replaced. With purge valve removed pressure did not go down.
4th I removed the line running from pump to autosampler, pressure didnt change. pressure is at 388 at this point with pump not running. I turn on pump set at 0.1ml/min 100%MeCN within 2 seconds shuts off at 418 bar. Is decreasing at about 1 bar/min.
5th shut down instrument, shut down chemstation, power cycle. No change.
Blockage seems to be somewhere in pump, but not sure what to look at next. Any ideas?