Pressure changes when injecting
Posted: Thu Jan 07, 2010 9:04 pm
We are seeing quite a bit of pressure fluctuation after each injection on our HPLC. For the first several minutes after each injection the pressure fluctuates about +- 10 bar and then it stabilizes. As we were troubleshooting and rinsing the instrument we noticed that the pressure increases about 100 bar if we change the injector valve from the inject position to the load position. This does not occur on our other instrument of the same brand.
Can anyone tell me what would be a normal fluctuation in pressure as the injector valve is changed? Does any one have any ideas on how to fix this? We've tried doing sample loop and buffer loop rinses with water and whole system rinses with IPA.
Thanks very much for any suggestions.