Packed Column - no analite peaks
Discussions about GC and other "gas phase" separation techniques.
5 posts
Page 1 of 1
I am very unfamiliar with packed columns. I am trying to run menthol at 100ppm in chloroform; I get solvent peaks but no menthol peak. Any idea why?
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Maybe 100ppm is to small for you to detect. Start with significantly higher concentrations of menthol, maybe 10,000ppm and work your way down.
Then you'd at least know what the menthol retention time is.
Then you'd at least know what the menthol retention time is.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
There are lots of different things you might have done (I and others in the past as well) which can give you these results.
The output of your amplifier may be set too low.
The temperature of your oven may be set too low.
The loading of your column may be to low so the menthol and solvent don't separate.
Your detector is not working properly.
Your sample may be splitting off before it reaches the detector (a leak).
Of course, with more clearly detailed information in your question it is more likely that someone will be able to give you the correct answer rather than guessing.
good luck,
Rod
The output of your amplifier may be set too low.
The temperature of your oven may be set too low.
The loading of your column may be to low so the menthol and solvent don't separate.
Your detector is not working properly.
Your sample may be splitting off before it reaches the detector (a leak).
Of course, with more clearly detailed information in your question it is more likely that someone will be able to give you the correct answer rather than guessing.
good luck,
Rod
-
- Posts: 70
- Joined: Fri Oct 14, 2005 1:15 am
I agree with Rod, in fact check that the GC is actually on, i came across 3 postdocs trying to get separation from a GC that was off! And NO, i am not pulling legs here. Fair Dinkum stuff!
Greg
Greg
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
I helped a PhD in a different department in the 1970s who had a giant leak in the (packed column) inlet end, so he couldn't even get a solvent peak. Once I noticed that, the fix was easy...
5 posts
Page 1 of 1
Who is online
In total there are 80 users online :: 0 registered, 0 hidden and 80 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 80 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 80 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science