Quantification methods

Posted: Wed Jan 06, 2010 4:00 am
by badrul
In pesticide residue analysis, which are the more appropriate quantification methods? standard in pure solvent calibration curve or matrix-matched calibration curve?
thank you..

Posted: Wed Jan 06, 2010 4:11 am
by Don_Hilton
I would suggest matrix matched calibration curves. Any compunds that may be lost on active sites in a GC benefit from matrix matched calibration.

Posted: Wed Jan 13, 2010 5:24 pm
by Jumpshooter
Suggest matrix-matching or the 'method of standard addition' to prep your primary STD calibrator. You will encounter far less problems when calculating your 'final analytical result'; because issues of different solvent density, %w/w concentrations will not have to be accounted for in your back end calculations.