Manometer connection for inlet pressure testing

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hello,
We are trying to connect a manometer to the split/splitless inlet to test for pressure decay and are having a very difficult time making the actual physical connection. Has anyone done this? If so, what sort of tubing/connectors/reducers have you used? Our GC is an Agilent 6890. Thanks!

You should be using either the pressure sensor that is built into the EPC version of the 6890 capillary inlet or the pressure guage that is built into the non-EPC version of the capillary inlet. There should be no need to connect a manometer.

Gasman

If for some reason you don't want to use the system pressure gauges/sensors as Gasman is suggesting, you can simply come up with one like that shown in the link below using a spare pressure gauge you have in your lab.

http://www.discoverysciences.com/product.aspx?id=2254

I would be really interested to hear why you want to do this :scratch:

Peter
Peter Apps
4 posts Page 1 of 1

Who is online

In total there are 52 users online :: 1 registered, 0 hidden and 51 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 51 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry