Our Thermo LTQ will not tune due to the mass of cal mix bouncing around as shown:
https://drive.google.com/file/d/1wsMEuq ... sp=sharing
The cal mix consists of caffeine, MRFA and Ultramark 1621. We have all green diagnostics and readbacks. I have even tried a different source from an instrument I know is working, and we observe the same. The instrument will pass the multipole and main RF calibrations.
Our settings were taken from the manual detailing the initial tuning:
Positive polarity
Spray voltage: 4.50 kV (though I have gone to the max with no change)
Aux and sweep gaes: 0
Sheath gas: 10
Capillary temp: 275 (I've gone to 350)
Capillary voltage: 13 (I've gone to 35)
Tube Lens Offset: 100 (I've gone across the entire range of +/- 140)
Max injection time: 200 ms (I tried the max of 8000 ms)
Our conversion dynode reads -15kV and both multiplers are at -1200 (I've ramped them to -2500 with no change)
Changing between +/- polarities shows the same results.
I even cleaned the multipliers following this document from Agilent and did see an increase in intensity: https://community.agilent.com/technical ... /files/654
It feels like it's something electrical but I'm not sure what it could be. At this point, I'm trying to find someone we could pay to send our top cover and coil box to as that's all I can think of. I've looked at the trap wiring based on the manual and it looks correct.
Any help/feedback would be greatly appreciated!