GC small volume samples - How low can I go?
Posted: Mon Nov 28, 2022 7:36 pm
Hi,
As part of an analytical method I am testing, I will be extracting analytes (at ultra-trace levels which have been deposited on a solid matrix) using various extraction solvents. These will be recovered, evaporated to dryness and will then be reconstituted in my analytical solvent ready for GC-MS.
Because my analytes are at very low levels, I would like to minimise reconstitution volume and wonder if anyone had any experience of small volume samples. I would like to go down to around 10µL (if that is practicable) and have bought a bunch of vial inserts (the ones with plastic feet) with a retention volume of 1µL which I guess gives me the best chance.
With this in mind, I would be really grateful if you might consider the following questions:
1. Is 10µL a feasible volume? Has anyone gone lower successfully?
2. Will the needle depth require adjusting given that the plastic springs push the inserts up fairly high?
3. Obviously auto-sampler 'sample wash' (where the sample is drawn up to remove residual solvent from the syringe) is not feasible... Should I also get rid of the auto-sampler bubble removal step (where the sample is drawn up and pumped) too?
I recognise that issue 3 is far from ideal and that my RSD will suffer, but on the plus side, the method is qualitative...
Many thanks for your opinions on this.
Kind Regards
TD2
As part of an analytical method I am testing, I will be extracting analytes (at ultra-trace levels which have been deposited on a solid matrix) using various extraction solvents. These will be recovered, evaporated to dryness and will then be reconstituted in my analytical solvent ready for GC-MS.
Because my analytes are at very low levels, I would like to minimise reconstitution volume and wonder if anyone had any experience of small volume samples. I would like to go down to around 10µL (if that is practicable) and have bought a bunch of vial inserts (the ones with plastic feet) with a retention volume of 1µL which I guess gives me the best chance.
With this in mind, I would be really grateful if you might consider the following questions:
1. Is 10µL a feasible volume? Has anyone gone lower successfully?
2. Will the needle depth require adjusting given that the plastic springs push the inserts up fairly high?
3. Obviously auto-sampler 'sample wash' (where the sample is drawn up to remove residual solvent from the syringe) is not feasible... Should I also get rid of the auto-sampler bubble removal step (where the sample is drawn up and pumped) too?
I recognise that issue 3 is far from ideal and that my RSD will suffer, but on the plus side, the method is qualitative...
Many thanks for your opinions on this.
Kind Regards
TD2