I'm doing the PM's/calibrations of about 20 HPLC's this summer. One of the tests is to run a gradient mixing to check the mixing of a binary pump system of shimadzu 10vp's.
The test is water, and .3% aqueous acetone. This is a corporate procedure so I can't change it, but I'm wondering if you guys see anything wrong with it.
What i'm finding is that EVERY single test so far (on about 5 different machines) is getting the same results: the acetone pump pumps about 1.5% too much at 50%, and .75% too much at 10% and 90%. The 50% fails the instruments and the only way I can make them pass is to change the alpha values on one of the pumps. Flow rates are perfect. Any ideas? Column is either a Nova-Pak C8 or C18