Manual Injection: Bigger carryover peaks & diluted samples
Posted: Sun Nov 06, 2022 1:44 pm
Hello all,
As the title says, I've been experiencing two separate issues pertinent to the manual injection method I'm using for analyzing my samples:
1. The carryover peaks recorded for blank injections is larger than the peaks recorded for the actual sample
2. Adding a "wash" injection to mitigate problem (1) dilutes the peaks recorded for the sample
I've narrowed down the source of problem to the glass syringe and/or Rheodyne 7725i sample injector but I'm unable to solve this problem which has resulted in excessive delay in my work. Does anyone have any idea regarding:
A. What is causing the carryover peaks to be larger than the sample peaks?
B. How can I solve this problem with the carryover peaks without replacing it with problem (2)?
For information, please find the pertinent information regarding the situation below:
1. Sample: Monosaccharide fermentation broth incl. organic acids & alcohols
2. HPLC: Shimadzu Prominence series
3. Column: RHM column (Pheonomenex)
4. Mobile Phase: H2SO4 (0.03N)
5. Column temperature: 40 degrees
6. Detector temperature: 40 degrees
7. Flow Rate: 0.6 mL/min
8. "Binary" gradient pumping same mobile phase through both pumps*
9. Sample loop volume: 20 uL
10. Injection volume: 150 uL
Notes:
* The "binary" gradient is being used to mitigate the problems with frequent pump failure caused by air bubbles
- The null injection does not produce any peaks, so I've ruled out the column as a source of the problem.
- I do wash the glass syringe with water and/or the mobile phase before every injection, & occasionally the sample injector too.
- Not sure if this issue is due to azeotrophy of the compounds with water and/or mobile phase
- Problem mostly observed with glucose and lactic acid.
- I've observed that lack of liquid draining out from the injector upon overfilling the sample loop
As the title says, I've been experiencing two separate issues pertinent to the manual injection method I'm using for analyzing my samples:
1. The carryover peaks recorded for blank injections is larger than the peaks recorded for the actual sample
2. Adding a "wash" injection to mitigate problem (1) dilutes the peaks recorded for the sample
I've narrowed down the source of problem to the glass syringe and/or Rheodyne 7725i sample injector but I'm unable to solve this problem which has resulted in excessive delay in my work. Does anyone have any idea regarding:
A. What is causing the carryover peaks to be larger than the sample peaks?
B. How can I solve this problem with the carryover peaks without replacing it with problem (2)?
For information, please find the pertinent information regarding the situation below:
1. Sample: Monosaccharide fermentation broth incl. organic acids & alcohols
2. HPLC: Shimadzu Prominence series
3. Column: RHM column (Pheonomenex)
4. Mobile Phase: H2SO4 (0.03N)
5. Column temperature: 40 degrees
6. Detector temperature: 40 degrees
7. Flow Rate: 0.6 mL/min
8. "Binary" gradient pumping same mobile phase through both pumps*
9. Sample loop volume: 20 uL
10. Injection volume: 150 uL
Notes:
* The "binary" gradient is being used to mitigate the problems with frequent pump failure caused by air bubbles
- The null injection does not produce any peaks, so I've ruled out the column as a source of the problem.
- I do wash the glass syringe with water and/or the mobile phase before every injection, & occasionally the sample injector too.
- Not sure if this issue is due to azeotrophy of the compounds with water and/or mobile phase
- Problem mostly observed with glucose and lactic acid.
- I've observed that lack of liquid draining out from the injector upon overfilling the sample loop