Autosystem XL gas chromatograph

Posted: Tue Nov 01, 2022 8:23 pm
by andrzej_w
Hi!
I'm trying to breath new life to our old Autosystem XL GC.
And of course I count on your help.
In particular I'm curious if anyone have a description of those DIP switches located on the right hand side of the chromatograph. Those located near RS232 and GPIB sockets.

With my best regards
Andrzej

Re: Autosystem XL gas chromatograph

Posted: Wed Nov 02, 2022 6:47 pm
by mburleson
Hi!
I'm trying to breath new life to our old Autosystem XL GC.
And of course I count on your help.
In particular I'm curious if anyone have a description of those DIP switches located on the right hand side of the chromatograph. Those located near RS232 and GPIB sockets.

With my best regards
Andrzej

While it isn't direct help, Perkin Elmer should be able to provide you with the manual to the instrument if you email them. I had to do this for the same system at my previous institution and for a PE 200 HPLC we were donated. I no longer have access to either or I'd send it your way.

Re: Autosystem XL gas chromatograph

Posted: Sat Jan 07, 2023 10:05 pm
by chromolover
hi andrzej, i have the same question, did you find out about the dip switches ? (i mailed PE maybe in august or september 2022, still waiting a reply)
and excuse my curiosity, but what software are you using with your PE XL ?

Re: Autosystem XL gas chromatograph

Posted: Mon Jan 09, 2023 10:58 pm
by S_G
https://resources.perkinelmer.com/corpo ... anager.pdf
Check p. 180.
Mine is set to DIP switch 6 = on for RS232. The rest of the communication can take place through USB-RS232 adapter.

For software, I'm using TotalChrom 6.3.3, but Thermo Chromeleon 7.3 works too.

Re: Autosystem XL gas chromatograph

Posted: Wed Jan 11, 2023 7:07 pm
by chromolover
thanks a lot, pdf is about Link600 interface device, but seems to match with the Autosystem XL dip switches

i would like to create a RS232 driver for Unichrom, so i'm desperately looking for the PE command list, any clue ?