UV detection is poor due to lack of chromophores and fluorescence detection requires pre or post colum derivatisation.
That's what the OPA is all about!
Pamela, the first question is "are you running an existing, validated method, or are you trying to develop your own procedure?"
If you are running an existing method, then it *should* work. Possible reasons for failure:
- the derivatization reaction with OPA is failing
- the derivatization is working, but the derivatized AA's are decomposing before chromatography
- the derivatized amino acids are not eluting from the column
- One or both of your detectors are malfunctioning.
If you are developing your own method, many of the same failures are possible, but troubleshooting will be more difficult, because you do not have a "known" response to compare.
If this were my problem, I would address the issues one at a time. For starters, work with only one amino acid!
1. Derivatization reaction. Remove the column and connect your injector to your detectors with a piece of capillary tubing (1 meter of 0.25mm id tubing should work) in place of the column. Prepare your sample per usual and inject. You should see a rapid "spike" response from the both detectors. If you do *not* see a response from the fluorescence detector, that means that either the detector is malfunctioning or that the derivatization reaction has failed. In that case, run the detector diagnostic tests as recommended by the manufacturer. If the detector checks out OK, then the problem is your derivatization reaction. I'm afraid I can't help on that, beyond suggesting the obvious: pH, temperature, reagent purity, etc.
2. If you *do* see a response with fluorescence, that tells you that the detector and derivatization reaction are working. Now prepare a fresh sample and repeat the injection every 10 minutes or so for several hours, while tracking the response. That information will tell you something about the stability of your derivative in solution. If it is decomposing quickly, then you will have to make sure that you prep a sample immediately before injection.
3. If sample stability is not an issue, you can re-install the column. Make an injection and run a simple full-range gradient:
A = your aqueous buffer + 5% acetonitrile
B = 100% acetonitrile
Flow = 1 mL/min
0 - 100% B in 20 minutes.
You should see a peak come out *somewhere*. It may be ugly, but it should be a peak. If you do *not* see any peaks, that suggests that either your buffer has been prepared incorrectly or your column is very, very dead.
4. If you *do* see a peak, that means that your derivatization, detectors, and column are OK. That suggests that your gradient conditions are bad. You did not specify what the % organic solvent was in your A & B solutions. For reversed-phase gradient HPLC to work, the "solvent strength" (approximately equal to the % organic solvent) should increase during the run. Identify what the % organic is when your peak elutes from the column. Now go back and look at your original method to check that you are going to at least that high a % organic. If not, that is your problem, and you will have to modify your method accordingly.
5. Once you have gone this far, you can add additional amino acids and figure out the best conditions for your separations.