PAH method for GC MS MS with PTV (CIS Gerstel)

Posted: Tue Nov 01, 2022 1:47 pm
by Hefe1995
Hello together,

i need to change an established PAH method. The reason is that i have to analyze four more PAHs. These new PAHs have a higher evaporation temperature and i cannot find them with the method we actually use.

- Benz[a]anthracene
- Benzo(b)fluoranthene
- Benzo[a]pyrene
- Chrysene
- Dibenz[a,h]anthracene
- Indeno[1,2,3-c,d)pyrene
- Benzo(k)fluoranthene
- Benzo[g,h,i]perylene

GC: Agilent 7890A
MS: 7000A Triple Quad
coupled with a CIS and TDU from Gerstel.

As i am new to GC MS MS (or the system we use) i am a little shy/careful to just change some of the parameters like source or cis temperature, cis injection speed and so on.

Does anyone already have an established method or know what i could change or try ?
I can provide any additional information :)

Thanks a lot

Re: PAH method for GC MS MS with PTV (CIS Gerstel)

Posted: Thu Nov 03, 2022 1:28 pm
by James_Ball
What are the concentrations of your lowest standard?

For large volume injection of those analytes I would not adjust the inlet parameters except possibly the upper temperature, but any of those will vaporize even with the inlet at 250C, 275C is a little better but you don't need anything drastic.

Using a 5 phase column (Rtx-5ms or HP5-MS ect) you will need to go to about 325C on the upper end of the oven program but they will all elute at those temperatures.

Re: PAH method for GC MS MS with PTV (CIS Gerstel)

Posted: Thu Nov 03, 2022 5:57 pm
by TCCrab
I've had really good luck with an Agilent DB17MS column.
Inlet temp at 280C, ramp from 75c to 300c.

TomC