viewing %RSD results empower 3
Posted: Wed Oct 26, 2022 1:35 pm
by pstaunton
hi,
is there a way of viewing your %rsd results for a run without having to go into a report?
Re: viewing %RSD results empower 3
Posted: Wed Oct 26, 2022 2:03 pm
by DR
results tab
Posted: Wed Oct 26, 2022 3:40 pm
by pstaunton
fantastic, i wonder is it a field that you have to set up yourself because im not seeing any columns related to %rsd under the results tab?
Posted: Wed Oct 26, 2022 6:04 pm
by DR
Assuming you are looking at injections or unprocessed results, you will have to do some online "processing" and add the %RSD column to your table beneath the chromatogram.
Posted: Thu Oct 27, 2022 9:22 am
by EmpowersBane
I dont think there is a "%RSD" column in the chromatogram view in Empower. You either add it to a column in a report table or you can create a custom field for Empower to calculate the %RSD and this value will populate when you generate a result set. For example, if you wanted the %RSD of amount in all your samples, you could create a formula like U%.%..%RSD(Amount) and add Summarize Custom Fields as the last line of the sample set, and it will then calculate the %RSD of amounts for all samples beginning with the label U...when you are looking at the result set.