Quadratic Curve vs. Linear Weighting?
Posted: Wed Oct 12, 2022 12:56 pm
Looking for a little help if possible.
I have a 6 point calibration that has lower accuracy at the low end (about 105-110% of true value). Verified by 6 independent calibration series' made on different days. Accuracy at my lowest point is always the same.
I have done an F-test with 6 independently prepared standards at the top, mid and bottom range of the calibration and shown the data is heteroscedastic and the sum relative error is the lowest with 1/amount2 weighting.
Out of interest I also decided to try a Quadratic fit with no weighting and it also gives much increased accuracy; even better than the linear weighting above for all curves and brings the low calibration point accuracy down to around 102%. It has also improved LLOQ accuracy from 115% to 101%.
Is there a standard way to choose which route I go with; quadratic fit or linear weighting? Detector is a D.A.D which I thought were always linear? Samples are derivatised...could this be a source of non-linearity?
Thanks for any help.