As well as the excellent advice from Merlin, a couple more points.
1. For fermentation monitoring, we used to get 5 year column life ( about 8 - 10,000 injections ) by diluting samples before microcentrifuging ( rather than filtering, which can get expensive with lots of samples, and also can lose some compounds , such as metabolites - if you are looking for them ).
If you are looking for trace components, you may not be able to dilute, and may need to filter or concentrate.
2. Temperature control is critical, we used to run the columns in a 35C temperature-controlled waterbath with 40C RI setting. The higher temperatures produce sharper peaks, shorter run times, and greatly reduce junk retention on column. Constant temperature and flow greatly enhances retention repeatability as well, we kept ours running 24/7.
3. Mobile phase is about the cheapest you can get, but use really good quality Milli-Q water and concentrated acid. Don't allow the column to stand with just water present ( if you switch from the current acid mobile phase ), wildlife can grow and quickly kill it.
4. Use a guard column - not just a prefilter, but a guard column, which is virtually mandatory if using a silver-based column, unless you really want to keep meeting the local column-selling representative.
Guard columns are especially important if samples are from research fermentations with unusual media. Large-scale defined media fermentations are much "cleaner" than small-scale research broths, which can produce diverse snot that is difficult to process and shortenens column life.
Another company that makes alternatives, and has some good free information, is Phenomenex with their Rezex range.
Please keep having fun,
Bruce Hamilton