which is the acceptance criterion of the peak purity? Is a value of 0.95 acceptable? Is there any official guideline about this?
thank you
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
An unreliable method of determining whether the separation method is sufficiently selective is to perform a spectral peak purity check. The absence of hidden peaks of impurities beneath the primary peak cannot be demonstrated by this purity check. The absence of any discernible spectral differences between the various components of the main peak (these differences are sought in the peak purity check) may be caused by low concentrations of impurities and the similarity of their spectra to the spectrum of the parent substance. I think this answer to the above question."Peak purity" (spectral) is a software/algorithm-dependent quantity. No official limits exist.