Hello my Chromatographical Friends,

Since a few days I have started working with Chromeleon 6.8 It takes some effort to make the transition from working with Atlas and ChromQuest to Chromeleon, but it seems I'm getting there.
There is just one thing I haven't been able to figure out. I usualy work with an external calibration sequence and I want to use this sequence for newer sequences with unknowns in them. What I do now is that once the sequence is finished I copy the standards + the raw data from the original calibration sequence in the unknown sequence. With that copied data I can quantify the peaks in the unknowns.
From the manual I understand that it is possible to use the external calibration sequence without copying the data, but how to do this? It has something to do with the 'Fixed'-calibration option in the QNT-method, but I don't know how.

Thanks!
Maurice

In the qnt-method set the calibration mode to 'fixed'. Thats on page 'general'. Then go to page 'calibration'. Select the line in the calibration table, right click on it and select 'append standard'. Now you can select a standard run.
If the page 'calibration is not visible, click in the main menu on 'view' and then on 'select sheets'.

Alex

In the qnt-method set the calibration mode to 'fixed'. Thats on page 'general'. Then go to page 'calibration'. Select the line in the calibration table, right click on it and select 'append standard'. Now you can select a standard run.
If the page 'calibration is not visible, click in the main menu on 'view' and then on 'select sheets'.

Alex
Thanks very much Alex! That worked really well.

Now for me to figure out how to make a summary report to my specs. :)

Maurice

it's also possible to create a sequence using the wizzard; select no standards and select the qnt from the calibrated sequence. The new sequence has a copy of the qnt with modus fixed.
4 posts Page 1 of 1

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry