noise on FID when connect second detector-HP5890

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Dear friends: we work on the University and have a 5890 GC, with a FID detector.
Few months ago, we've build a second detector (electroantennogram) and connected it to our system. Since we connected the power contacts (heater and PT-100) we are having randomic noise in our FID signal (negative peaks, etc). We are not using the signal of the 2nd detector yet, so maybe some interference between the power cables and the FID' signal ones. We were carefully avoiding crossing between power and signal cables, so what's happening?
Does anyone have an idea?

Thank you in advance. Sincerely yours

DL

I've seen this problem in the past. The problem is noise from the heater being picked up by the signal cable of the FID. Try wrapping all the cables in aluminum foil to minimize the interference.
Thanks Bob, I'll try that.
regards

DL
The detector's body and the chasis must be isolted because of grown problems, that's all!
Thanks anyway
4 posts Page 1 of 1

Who is online

In total there are 8 users online :: 3 registered, 0 hidden and 5 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot], Georguslat, Google [Bot] and 5 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry