determination of viny chloride using ITEX and GC-MS
Posted: Tue Dec 08, 2009 1:41 pm
Hello!
I am trying to develop a method for the determination of vinyl chloride and other volatile compounds (Freon 113, MTBE, cis-DCE, TCE and PCE) by using the in-tube extraction with GC-MS.
My system is a Thermo Scientific DSC II, Trace GC, split/splitless injector.
My column is an Agilent DB-VRX 60m x 0.32mm x 1.80um.
The optimal temperature program that I am using for the determination of VOCs except VC is: 40oC for 2min, 15oC/min to 180oC(2min), column flow rate:1ml/min. Temperature of the injector 250oC, transfer line 235oC, direct coupling to the MS, EI mode, ion source at 200oC, scan/s:1, scan range 35-190.
I am using a splitless mode with a liner of 1mm, to have the greatest sensitivity.
the parameters of the PAL are:
Syringe volume: 2.5ml/ TENAX GR
PAL conditions: Incubation temperature: 70oC
Incubation time: 5min
Agitation Speed: 500rpm
Syringe temperature: 60oC
Extraction volume: 1000uL
Extraction strokes: 50
Extraction speed: 200ul/s
Desorption temperature: 250oC
Desorption speed: 100ul/s
Flush time: 2min
Pullup delay: 5000ms
Trap cleaning temp: 250oC
This program works great for the 5 of the 6 compounds, when I am adding the vinly chloride, the peak is really broadening, I know that it happens because of the lack of cryofocusing, and the backflush is huge. So, I tried to put a retention gap to concentrate the VC, but I always have a leak in my system, even if that I tried different connections and specific pressfit. Also, I tried to make a flow program where, I increased the flow rate to 2ml/min for 0.5min and then with the highest ramp to go back to 1 ml/min, but the VC still has the same broadening, and my other peaks had a little bit changed.
I played too, with temperature column programs, but it didn't work.
Next step,I think it is to use split mode, but I would like to have the greatest sensitivity.
Is there anyone that has any other suggestions, except to put on a cryofocusing?
I would be thankful with any answer.
bsxenia
Flush time: 2min
bsxenia