no MS trace signal on LC/MS (waters ZQ)

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Greetings,

I run one of our drug molecules on Waters LC/MS, and did not have any MS trace signals even with UV trace of about 1.6 V signal.

moble A: 0.05% TFA
moble B: ACN
MS + mode

Please let me know possible reasons for this lost of MS signals. Is it because some molecules are extremely hard to ionize? I cleaned cone and cone gas nozzle, but still did not get MS signals.

Thanks a lot

Does the MS still work for other molecules?

So this is a triple quad instrument? What scanning mode are you using (Q1, MRM, SIM)? You should at least see some noise, especially in single MS mode (Q1 scan). What about if you inject sample directly into the MS with no HPLC separation?

problem solved because of clogged capillary, so it did not work for other molecules either.

thanks everyone.
4 posts Page 1 of 1

Who is online

In total there are 29 users online :: 1 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry