Is the a guideline on plate count for system suitability.

CDER recommends >2000

If I have a plate count of the first analyte of 1700, is the system suitable.

Jay

2000 is a recommendation, not a requirement.

If this is a validated method and its system suitability criteria call for 2000 plates, then no, your system is not suitable.

If you are developing and validating a method, then you should put in appropriate system suitability parameters. If it is a simple assay with no interfering peaks, and you can prove that the method meets your criteria for accuracy, precision, etc., then 1700 may be adequate. But, because your plate count is lower than the "recommendation", you should expect some difficulty from reviewers!
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 14 users online :: 0 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry