Hello,
I have two questions.
I'm making a GC analysis using an autosampler and I'm calculating the CV%. After 9 injection I obtain an area CV% under 1% .
My first question is: is it a good CV%?
My second question is: I obtain height CV% also of 2-3%, why this difference between area and height CV%?
Thanks

Greta

< 1% is what I would expect (usually from 5- 6 injections).

I would generally expect area RSDs to be less than peak height (but it's been so long since I used a method based on peak height i've forgotten why). How is the RSD for the retention time? small fluctuations in carrier gas flow could lead to wider peaks which may have a minmal effect on the area but affect height.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.

Normally in chromatography, the peak hight is not effected by flow, but the area is directly dependent on it. The lower precision with peak hight is probably due to the lower intrinsic averaging of noise in peak hight as compared to area. (The integrators which I know take the highst point on the apex for determining peak hight, so no noise averaging there).
To put HW Mueller's answer another way - the GC software only records a data point every so often. Maybe 20 times per second. For a fast, narrow peak, you may not always record a measurement exactly at the top of the peak, so there will be some error (noise) introduced. (If you greatly enlarge the time axis on your plot, you can see the peaks are actually a bit jagged and not nice and smooth.

For peak area, you are recording 20, 30, or more points across the entire peak and then integrating. This is less dependent on the exact location of the points across the peak (i.e. it is signal averaging).

Your %CV looks good to me - to get better can be difficult and depends on what you are analyzing.
10 Agilent 6890, 1 6850
TCD, FID, & MSD

Thanks to all guys!
And so my question is: can I do not worry about the height and refer only to the area?

Daniela

The majority of GC quantitative work is done using area instead of height. Most regulated methods use area and not height. There should be no problem using area for your methods.
6 posts Page 1 of 1

Who is online

In total there are 337 users online :: 1 registered, 0 hidden and 336 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 336 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry