Calibration Report Issue Empower 3
Posted: Tue Aug 16, 2022 2:47 pm
by HerbalChem
My company is trying to make a single report method for every test that we perform. When trying to report a calibration curve in Empower 3 I choose the result set and open publisher with the established report method the system wants to generate a calibration graph and table for every injection from the run. I've tried only selecting the results that pertain to the cal curve and I still run into the same problem. I've played with every setting I could think of and it doesn't seem to resolve the issue. Is there something I'm missing?
Re: Calibration Report Issue Empower 3
Posted: Wed Aug 17, 2022 6:53 pm
by DR
Method properties: Summary by All
Posted: Mon Aug 22, 2022 4:05 pm
by HerbalChem
I've tried Summery by All and it still gives the same duplicates just in a single report.
Posted: Mon Aug 22, 2022 6:59 pm
by DR
Posted: Tue Aug 23, 2022 9:17 am
by EmpowersBane
Can you not use a data filtering condition in your report table and only use one sample? For example, say you run 10 samples 1A to 10A, filter by SampleName = 10A, that way only one cal curve graph and table is shown but its still relevant for all the samples, presuming all samples are quantitated off this curve.
Posted: Wed Aug 24, 2022 9:55 pm
by HerbalChem
That seems to have worked. I'll try it out with different result sets to verify. Thank you!