Peark area is 90% less when switiching from 5890 to 6890.
Discussions about GC and other "gas phase" separation techniques.
6 posts
Page 1 of 1
When injecting a sample on a both a 5890 and 6890 we are seeing only 10% of the sample on the 6890 when compared to the 5890. The column and liner are the same on both instruments. The split and column flow are the same. Even the syringe used to inject the sample as well as the operater are the same. We have done this test on two different 5890 GCs and got similar results. Why would the 6890 show such a low conc? We have done troubleshooting for leaks, changed to a new column, and nothing changes the results. The signal value is lower on the 6890 when compared to the 5890. Would this have an effect? Any suggestions or comments would be appreciated.
-
- Posts: 29
- Joined: Fri Jun 20, 2008 7:59 pm
What kind of detector is on each of the instruments?
Wasson-ECE Instrumentation
http://www.Wasson-ECE.com
http://www.Wasson-ECE.com
-
- Posts: 10
- Joined: Fri Mar 07, 2008 2:48 pm
They are both FID.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Could you elaborate a bit on the problem? Do you mean that the calibration plots look OK on both instruments and calculated sample concentrations are down 90% on one? Or do you mean that *all* responses (calibrators and samples) are lower?seeing only 10% of the sample
If it's the latter, check down around the bottom of you calibration curve and look at the signal/noise ratio. It's possible that you may just be seeing a voltage scale difference. If your signal and noise are both 10%, your sensitivity is actually the same.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 418
- Joined: Thu Sep 25, 2008 3:40 am
How about peakshapes and retention times?When injecting a sample on a both a 5890 and 6890 we are seeing only 10% of the sample on the 6890 when compared to the 5890. The column and liner are the same on both instruments. The split and column flow are the same. Even the syringe used to inject the sample as well as the operater are the same. We have done this test on two different 5890 GCs and got similar results. Why would the 6890 show such a low conc? We have done troubleshooting for leaks, changed to a new column, and nothing changes the results. The signal value is lower on the 6890 when compared to the 5890. Would this have an effect? Any suggestions or comments would be appreciated.
Have you measured the split ratio manually?
Also please check the column installation at both ends.
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
How are you measuring the output of each GC. Are you using the analog output to an integrator or A/D device. If yes, the fact that you get a signal 10% of the 5890 suggests to me that you could have a difference in the Range settings. This is the only control on the 5890/6890 that changes things in steps of 10.
Gasman
Gasman
6 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science