Advertisement

CTC Pal headspace problem

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hi all,

i currently use an Agilent 7890 GC coupled with a 5975MSD and a CTC pal autosampler. i need to develop a method where i use a different vial size (i.e height).

Could somebody tell me how to adjust this setting please?

Many Thanks

Steve

wrighty,

I don't have a CTC but a MPS (CTC modified and sold by Gerstel). They are relatively similar.
You have up load the objects (i.e. sample trays, syringe type, vial type, etc..) with a software called PALoader that should be present somewhere on your pc. Once you have done that you can start to modify the settings using the remote attached to your CTC.

Regards,
bhuvfe
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry