Chemstation B.02.00.589 Error Message

Posted: Thu Nov 26, 2009 3:58 pm
by wrighty
Hi all,

just a quick question regarding an error message i am receiving when i try to run a sequence using this version of chemstation on a GC/MS system (Agilent 7890 coupled with 5975MS).

Following a column change (trim) i now get the following message:

7006-222,"({SP00} PARM EMUL,3278.223509;): Parameter emul value out of range [0.000,3000.000]"

I have re-calibrated the column settings but i cant figure out how this could affect the software in this manner.

Any help or advice would be greatly appreciated.

Cheers

Steve

PS the system passes its tune with no problems

Posted: Fri Nov 27, 2009 9:56 am
by Garlarg
Look around here for error messages:

[/url]http://www.chem.agilent.com/en-US/Support/Downloads/Patches/Pages/default.aspx[url]

You will need the product registration numbers.[/url]

Posted: Mon Nov 30, 2009 8:42 am
by wrighty
Thanks Garlag,

THe problem was traced to a dirty source. we belive the cause of this was due to an air leak when the column was re-installed. After giving the source a clean it is working as good as new now.

Thanks