lmh » Fri Nov 27, 2009 9:25 am
I think your approach is certainly reasonable. It's not perfect, but very few LOD measurements are going to be perfect!
In an ideal world you would do your LOD on a spiked blank matrix; you would use a matrix that has all the problems and none of the analyte expected in a real sample, and spike it with known amounts to make the low calibration curve for the LOD calculations. The difficulty is presumably that (1) you have no such blank matrix available, and (2) you're concerned that every sample matrix is different.
Running the LOD on a clean standard and scaling it according to the internal standard is certainly seems a reasonable approximation to what's going on in a real sample. My main reservation would be if you have peaks in the real samples that nearly coelute, or background ions/signal that appears in real samples but not in sequences consisting only of clean standards. These could increase the difficulties your integrator will have in quantifying the real peak, and could increase the variability of the real peak's area, which would invalidate your LOD. On the other hand, (a) you will probably be able to spot problems like this very easily by eye, and (b) if you get a sample that is apparently nothing, you can always follow JGK's very good advice and spike it with your expected LOD to see if you can indeed detect your LOD (good practise anyway).
Good luck.