column very high back pressure

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi All,

I amMr Khalid Rauf, a PhD scholar working on neurotransmitters, in distinct rat brain tissues, like, frontal cortex,striatum,etc using HPLC, with coulchem III, detector using MD-150 column.i use phosphate buffer 90:10 ACN, ph 3 and 0.6 ml flow rate.I experience more than3500psi pressure at 0.6 ml.How ca i clean my column .

waiting

sincerely yours
Khalid Rauf
University of Peshawar
Pakistan

You didn't specify the diameter. While I've never worked with that specific column, 3500 psi for 0.6 mL/min through a 150 x 3mm column sounds a bit high, but not unreasonable. If you're using the 1m ID column 3500 psi would be quite low.

First check to make sure there is no blockage elsewhere in the system. Then check with Dionex (ESA) confirm that back-flushing the column is OK. If so, then turn the column around and run it in reverse for a half-hour or so.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 8 users online :: 1 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry