Hello

I want to develop a method to detect and quantify ppm levels of ethanol (and aceton as a standard) in aqueous buffers

The system that we use is: Poraplot Q (5m), trace-GC and injection-needle 10µl (it's the smallest size possible)... the maximum split is 250

It's my first time doin' GC, so some suggestions (temperature and flow programme) would be very helpfull :)


Thanks!

Tracer,

This sounds like a student project. Perhaps you should have posted this in that section of the forum.

An isothermal oven should be adequate for the separation of ethanol and acetone, anywhere between 60°C and 200°C should work.

This issue is getting enough sample onto the column to detect 1 ppm.

A solid-phase microextraction fiber needle to sample your buffers should work fine for your needs.

Good luck on your homework.

best wishes,

Rodney George
consultant

I believe that Poraplot Q is sensitive to water and your samples will not run well.

Poraplot Q will be problematic with Aq samples.

A Thames Restek RTX-1 will do it.

A 30m column will give good seperation.

Headspace :wink: No sampler ? You mention trace-gc ? Thermofisher ?? If so get one of the special water injection liners 453-003-20, inject 1ul from your 10ul syringe with a low split ratio.

Oh well, I'm using PLOT Q for Water, IPA and Ethanol, so I think your system would be fine.

here's how I set it up
Injector 200C, 10:1 Spilt
Oven: 170 for 5 mins
Detector: 250C, Hydrogen 60, Air 300, Make up He 3

Hope it helps.

I analyzed traced water in an organic solvent and with Plot Q. I am not sure the reverse could work: trace ethanol in water.

Thanks for your comments.

We are now working with phosphate buffer with ethanol (100mM-0mM) and aceton as standard (10mM). 150°C-250°C at 20°C/min. Works fine, however, we have a peculiar problem:

When we set up a calibration curve (Area ethanol/Area aceton) as a function of the concentration ethanol, we obtain a rico x

When we then increase the pH of our buffer system (phosphate pH 6.5) to pH 10-11, we obtain a rico 1.5x (linearity remains perfect)


So what can be the explanation?

(Does the increased pH (injection volume only 1.2µl) damage the GC-column?)

If I understand your post correctly, you adjusted the pH of your phosphate buffer to a basic pH.

What kind of phosphate buffer are you using? sodium?

What ion are you adding to change the pH ?

What is the meaning of your abbreviation of RICO ?

Rodney George

If I understand your post correctly, you adjusted the pH of your phosphate buffer to a basic pH.

What kind of phosphate buffer are you using? sodium?

What ion are you adding to change the pH ?

What is the meaning of your abbreviation of RICO ?

Rodney George
sodium indeed... the pH was increased by adding some µls concentrated NaOH (10M) to the buffer... (alkalinization is needed to stop a reaction which would otherwise continue in the vial during analysis)

with rico i mean slope of the curve... sorry for the terminology

best regards
10 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry