Could you help me to troubleshoot this?
After downloading the method, pump rated at 0.7 ml/min, I noticed the pressure doesn't rise (0 - 2 psi), which may indicate the existence of a leak. When powered on, the run LED is amber (it should be green), but it flashes green while equilibrating the column but when it is time to perform a run, the run LED turns amber and the autosampler state says 'waiting for pump'.
The main window (PDA detector) status bar says: waiting for trigger. And nothing happens then.
Question: Despite the existence of a leak, the run LED should be flashing green, right?
Question: Why can't the run LED be green (without flashing) at the start-up, meaning it is in the ready state?
The manual says if the run LED is ambar, there may be a few motives:
1 - The LC pump is not communicating with the data system PC;
2 - the pump’s motor is stopped;
3 - the pistons are homing, or
4 - the pump is stabilizing (except during a run).
Any help provided will be very much appreciated.
Thanks.