SCD

I have a SCD equipment with a chromatograph.
But, I have a problem, in the first injection, the peaks appears normally, but from the second, the peaks desappear.

Do you have a theory about this?

Thanks

what are the peaks?

Sulfur compounds, like H2S, COS, methylmercaptane.

What are you using to inject sample? A gas tight syringe, valves, what kind of inlet? What carrier gas?

I´m using a GSV, but before the valve, I'm using a inox tubing.
The carrier is Helium, but I only pass the gas sample through the tubing and valve, without carrier.

lynoguchi,

How old is your SCD? The older SCD's had a problem of reactive furnace tubes that would basically start to gobble up sulfur compounds from the very start.

Best regards.

The SCD is new. It was installed 2 months ago.
But I think the problem is the ceramic tube for pyro.
Maybe I need to change it?

lynoguchi,

Based on my limited knowledge of SCD, I would suggest replacing the tube as the starting point. Unfortunately, I do not have hands on with this detector although I hope that is about to change.

Best regards.
8 posts Page 1 of 1

Who is online

In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry