Okay, Bruce, what *exactly* do we have to do to "infer" accuracy?
accuracy may be inferred once precision, linearity and specificity have been established.
I've always suggested that recovery from a spiked matrix was the most meaningful surrogate for accuracy. And that when no blank matrix is available, that linearity and precision of standard additions would have to "make do". But danko makes a good point: how do we justify the leap from "it's precise" to "it's accurate"?
Refer the ICH guidance...
Section 1 discusses requirements for Specificity.
Section 2 discusses requirements for Linearity
Section 5 discusses requirements for Precision
The extract that I provided was intended to indicate that Linearity and Precision ( as listed by the original poster and others ) were inadequate to infer accuracy, you also have to demonstrate specificity.
I was not suggesting that experimental determinations are inferior, just that people seem to be forgetting about the essential specificity requirement that complements the precision and linear requirements to obtain what used be known as "trueness". If the analysis is precise, linear, and specific, is it not reasonable to infer it's accurate?.
The Guidance also clearly points out the need for independent procedures when available materials are inadequate.
Perhaps I should have posted more of the Accuracy section, so here is...
[ Begin extract ]
4.1. Assay
4.1.1 Drug Substance
Several methods of determining accuracy are available:
a) application of an analytical procedure to an analyte of known purity (e.g. reference material);
b) comparison of the results of the proposed analytical procedure with those of a second well-characterized procedure, the accuracy of which is stated and/or defined (independent procedure, see 1.2.);
c) accuracy may be inferred once precision, linearity and specificity have been established.
4.1.2 Drug Product
Several methods for determining accuracy are available:
a) application of the analytical procedure to synthetic mixtures of the drug product components to which known quantities of the drug substance to be analysed have been added;
b) in cases where it is impossible to obtain samples of all drug product components , it may be acceptable either to add known quantities of the analyte to the drug product or to compare the results obtained from a second, well characterized procedure, the accuracy of which is stated and/or defined (independent procedure, see 1.2.);
[ End extract ]
Please keep having fun,
Bruce Hamilton